[evlatests] On-Line Averaging is working.
Rick Perley
rperley at nrao.edu
Mon Dec 5 17:45:14 EST 2011
Tests were run today to see if the on-line averaging is working again.
Short answer: All seems fine. We can move ahead with regular
observing.
Details:
There are four databases, all centered at 2627 MHz.
1) 128 channels of 500 kHz width each, averaging = 1 second.
2) 64 channels of 1000 kHz each, averaging = 1 second.
3) 128 channels of 500 kHz width, averaging = 5 second.
4) 64 channels of 1000 kHz each, averaging = 5 second.
A quick calibration was done on each, using 3C286 as reference. The
targets were J1800+7828 (dot calibrator), and a noise field 10 minutes
of time away.
The rms noise (as determined using histogram widths, in the
cross-hand channels), with Hanning smoothing applied:
1) 263 mJy
2) 185 mJy
3) 127 mJy
4) 87 mJy.
The result from (1) predicts 118 mJy for (3). The observed is 127
mJy -- a little high.
The result from (2) predicts 83 mJy for (4). The observed is 87
mJy, also a bit high.
The small discrepancies between predicted and actual are likely due
to single antenna effects -- I made no attempt to remove edge effects
(this will preferentially affect the 5 second data, I should think), or
bad antennas (other than ea23LCP, which is very noisy).
I conclude that on-line averaging is working as expected.