[evlatests] Stellar high-frequency data
Michael Rupen
mrupen at nrao.edu
Sat Jun 26 15:32:22 EDT 2010
> Was your test able to say anything about phase transfer at high
> frequencies?
>
* It is unclear whether ref.ptg. is being applied -- Rick was looking
into that with recent data and found confusing results. I'm
assuming for now that this reflects windy weather rather than
something fundamentally wrong in the system; Ken might comment
on his recent pointing results which I think looked OK.
* http://www.aoc.nrao.edu/~mrupen/TestPlots/CptgCKobs_vplot.pdf
shows the results of a 32 minute run involving C-band pointing
followed by K and C observations with pointing applied. Greenish is C;
reddish is K. Interesting bits at K band are:
- phase jumps on 4, 8, 10, 11
- loss of fringes for some antennas/IFs for some scans
(usually all scans following a refptg but I've seen exceptions in
other runs)
- ea20 has the pointing problem Ken described and should simply be flagged.
- there are few to 5% amp jumps on several antennas/IFs
From these data it looks like you're basically OK if you calibrate
amp/phase after every refptg scan. Note however that occasional
scan-to-scan phase jumps are seen even just sitting at C band.
* We did a similar C-ptg + Ka/C-obs run which Vivek says looked quite
similar. Most of the remaining OSRO projects involve Ka band.
* There are indications that using X-band leads to more phase/amp jumps.
Based on this I believe we can observe high-frequency OSRO so long as
observers (1) point up at C band, (2) calibrate carefully after each
ref.ptg. scan (good practice in any case), (3) keep an eye out for full
and partial loss of coherence for individual antennas/IFs.
Of course the weather is making this mostly moot at the moment but we
may get some observations in Sunday night.