[evlatests] Alternating 'Extended OSRO' and 'RSRO' modes
Rick Perley
rperley at nrao.edu
Thu Dec 16 22:01:08 EST 2010
I discovered recently that the OPT now gives us the ability to
configure the correlator in ways not accessible to me in the past (so
far as I knew). I tried a test this afternoon where I set up the
correlator in a 'RSRO' mode with 1 MHz channel width, and 64 MHz subband
width -- this allows us to cover all of L-band, with full polarization,
and no 'overlaps' between A/C and B/D. The A/C half was set to cover
the lower half of L-band, the B/D half covered the higher half. I then
inserted this special L-band mode into my standard 'all bands as fast as
possible' sequence, which uses 'Extended OSRO' mode: 2 MHz wide
channels, covering a total of 2 GHz of bandwidth. The durations for all
eight bands (L-band in the 'RSRO' mode described above, and all other
bands in the standard 'Ext. OSRO' mode) were kept at shortest possible
values, giving 30 seconds of good on-source data. The cycle was run
three times.
The experiment worked perfectly. The 'RSRO' mode setup seemlessly
covers 1024 MHz of bandwidth, from 988 through 2012 MHz. All L-band
phases and amplitudes join seemlessly, except of course at the halfway
point (where A/C and B/D IFs are adjacent). The time lost between
recofigurating between these modes was negligible.
I'll next try dialing down the subband width, to get to 125 kHz
channelwidth -- the agreed-upon resolution for the 'full spectral sweep'
experiment.
More information about the evlatests