[evlatests] Some Good News!
Rick Perley
rperley at nrao.edu
Sun Jun 14 13:27:42 EDT 2009
Michael took an hour of data on our favorite northern naked source,
0217+734.
Differences from former tests are (so far as I know)...
1) X-band, rather than C-band.
2) 2 IFs, (sub-bands) rather than 4.
3) Point source was offset from phase center by a small amount
(about 0.3 arcseconds, I think).
There is a world of difference between this database, and all the
others taken this month ...
1) There are *NO* 180 degree phase jumps. All data are completely
seamless in amplitude and phase.
2) Except for the initial couple of minutes, there are no bad data.
I cannot recall ever seeing such lovely data ...
3) Histograms of the single-channel distributions (both real and
imaginary) show absolutely perfect gaussians, with 1-sigma widths of
0.655 Jy -- a bit better than what I expected.
3) Initial images showed our 'lumpy-bumpy' problem. But read on ...
To uncover the origin of these lumps/bumps, I SPLITed the data into
a single 961-channel average (to increase SNR) and plotted the amplitude
and phases for all baselines and times. The amplitudes looked just
fine, but the phases showed a single baseline with an oscillatory
pattern which persisted throughout the entire hour of observation. The
pattern frequency decreased throughout, reaching a zero at about 14:50
IAT, at a U-spacing of -2.1 Klambda -- i.e., this problem has nothing to
do with a steady external stationary signal (which must maximize when
the source fringe rate goes to zero -- at U = 0).
The evil baseline is 2 x 18. The amplitudes look fine, but the
phases are oscillating by +/- 2.5 degrees. Both IFs (sub-bands) behave
identically (except that IF#2's phase oscillations are inverted w.r.t.
#1).
I then flagged the evil baseline, and re-calibrated. BLCAL was run
(single solution for entire hour). The baseline corrections are
*extraordinarily* small -- typically 1 part in 10^5, consistent with
zero, I believe.
The single-IF (#1) map which resulted from this is almost perfect.
The rms (it doesn't matter where it is measured, the background is
completely flat) is 55 microJy -- this is *exactly* what it should be
based on the single-channel histogram width reduced by the square root
of the number of points (150602) and channels (916). The resulting
formal DR is 78200.
There is a barely detectable waviness seen beneath the noise. As
this is hardly attenuated at all throughout the image, it appears these
could be a very few unflagged points which escaped by (almost
non-existent) flagging.
So! Why is this X-band test so good, while the preceding C-band
ones so bad?
And -- we still have a single very bad baseline. What is wrong with
it?
Finally -- other than that bad baseline, the corrrelator looks to me
to be behaving precisely right.
I'll do more checks tomorrow. It's Sunday morning, and my garden is
calling me home ...
More information about the evlatests
mailing list