[evlatests] Polarization Results from 3C273/3C286 database
Rick Perley
rperley at nrao.edu
Thu Feb 21 17:03:06 EST 2008
I've looked at the polarization results from the long observation of
3C273 and 3C286 at X-band -- and there are some very curious things
going on ...
The observation was about 7 hours in duration, mostly on 3C273,
taken in polarization spectral-line mode: BW = 12.5, Nchan = 16 per
correlation. After careful flagging, and generated a 'channel 0'
single-frequency database by vector summation of channels 3 through 10.
Time averaging was left at 3.3 seconds, bandpass solutions were applied.
This database was then calibrated for polarization by standard
means. I used 3C286 (essentially unresolved in polarization and total
intensity), to generate the solutions, and rotated the phase frame to
ensure the polarization position angle was at 33 degrees -- the 'known'
absolute value. There were 7 observations of 3C286, five before
meridian transit, and two after.
An image of 3C286, in Stokes Q and U, showed good images, but much
noisier than expected. It was quickly found that antenna 14 showed
higher polarization visibilities than it should. Removing this antenna
improved the images, but still the noise is much higher than it should
be -- and is higher than in Stokes 'I'.
More careful review showed that both antennas 14 and 15 -- even
after polarization calibration -- give large offsets in polarization
amplitude and phase, against all other antennas. The effects are
particularly noticeable in phase: In Stokes 'Q' (correct amplitude is
235 mJy, and phase should be zero), well-behaved antennas show
amplitudes deviating not more than 10 mJy (these are scan averages), and
5 degrees. Antennas 14 and 15 give amplitudes up to 350 mJy -- note
that the error is always positive -- there are no baselines against 14
or 15 with too low a polarized flux! In phase, the patterns are most
interesting -- values of + or - 20 degrees are seen on nearly all
antennas. There are no values in the +/- 5 degree range -- it's always
near +/-20! The amplitude offsets are nearly constant in time, the
phases are not -- becoming much larger near meridian transit. The sign
of the phase offset is correlated perfectly with DCS number -- positive
offsets have a lower DCS, negative one are higher. This is caused by
the VLA's phase sign convention -- there is no deeper physical meaning.
Although antennas 14 and 15 (DCS numbers 16 and 13) are the worst
offenders by far, others should similar patterns, particularly antennas
7, 16, 19, 22 23 and 24, and mostly notably near meridian transit.
These troubles are clearly seen in the observations of 3C273 as
well, although because this source is heavily resolved in Stokes Q and U
(and the powerful point source in stokes 'I' is reduced to a few hundred
mJy in polarization), the Q and U visibilities are modulated
considerably by the jet's visibility. But despite this, it's obvious
that baselines to 14 and 15 are not correct in the cross-hand
visibilities.
I have no idea of what is wrong here -- but clearly the data and the
model used to separate antenna from source polarization are not in
agreement. This is not an EVLA issue -- both VLA and EVLA antennas are
equally in error.
More information about the evlatests
mailing list