[evlatests] 12.5 MHz spectral noise, and self-test
Rick Perley
rperley at nrao.edu
Wed Apr 23 19:28:19 EDT 2008
I ran a test this afternoon to see if the presence of self-test has
any effect on the noise histograms of data taken in 12.5 MHz spectral
line modes.
Answer: It does -- and the difference entirely accounts of the high
noise seen in the tests of recent weeks.
Details:
I ran short observations of 3C147 in 50-MHz continuum, and 12.5 MHz
spectra line in modes PA, PB and 4. The observations were run twice,
once in standard mode, and once with self-test turned off.
I have reduced only the PA mode data thus far -- the result is very
clear. With self-test turned on, the parallel-hand histograms of the
imaginary part of the visibility (for calibrated data) have a width of
1.0 Jy. (Channel width of 780 kHz, averaging time of 1.66 seconds).
With self-test turned off, the width of the histogram was divided in
half -- to 0.47 Jy. This is seen in both VLA-VLA baselines, and
EVLA-EVLA baselines. The effect can easily be seen even in the noise of
the CALIB solutions.
More information about the evlatests