[evlatests] Still more on spectral line noise
Rick Perley
rperley at nrao.edu
Wed Apr 16 22:44:40 EDT 2008
More tests were run this afternoon -- same setup as before
(calibrator, plus blank sky). This time, I observed in modes:
Continuum, PA, PB, 4, 2AC, 2BD, 1A, 1B, 1C, and 1D, all at each of three
bandwidths: 25, 12.5 and 6.25 MHz. That's 27 different setups.
In short, only the observations taken with 12.5 MHz BW, in spectral
line, give discrepant results. But the higher rms seen at 12.5 MHz BW
is not the same for all modes. I present a table, below, showing the
SEFD for each setup.
Mode BW=25 BW=12 BW = 6
-----------------------------------------------------
Cont. 340 340 325
PA 380 630/370 350
PB 380 580 380
4 370 580 330
2AC 340 430 320
2BD 340 470 350
1A 360 520 360
1B 340 405 380
1C 350 420 380
1D 350 660 340
----------------------------------------------------
Overall, the values are higher t(by ~10%) than the morning
observations -- but the cause is very likely the winds, which were near
stow limits throughout.
I can not think of any more useful tests. Perhaps somebody else
can. Better -- perhaps somebody can explain why the use of spectral
line modes with 12.5 MHz BW causes an apparent loss of sensitivity by up
to a factor of two. Note that for all observations, the data were
loaded as correlation coefficients, and for all datasets, the correction
required to convert these to flux density were the same. In other
words, the apparent loss of sensitivity seen in the noisier maps and
wider histograms for the 12.5 MHz data were *not* seen in the
correlation coefficients. How can this be?
More information about the evlatests