It looks to me that we are not turning down the requantizer gain enough
for P band. Our standard setup has a gain of 30 for the high bands,
where we have much of a GHz of bandwidth into the digitizer. At P band,
effectively all of the power from the digitizer comes out of the
deformatter, whereas at the high bands, the FIR filter in the deformatter
removes about 1000 MHz/ 64 MHz of the power. Therefore, the gain of the
requantizer should be turned down by a factor of 4, to about 8. It is
currently set to 15.
4 Band and P band are in the same digitizer band. We are attempting
to set them to approximately the same spectral power density (or maybe
a little higher for 4 Band). So the power coming from the FIR filter
in the two cases, P and 4 bands, is roughtly proportional to the
respective bandwidths, 40 MHz for P and 1.7 MHz for 4. However, the
same argument I just used above also applies to the analog filters in
the T5 module - they will turn the gain up by about the same factor,
so the requantizer gain can remain the same at 4 band as it is at P
band. It is currently set to 5.
The A & C IFs are leveled according to the detector readings. This should
produce the same RMS on the link as we see at the high bands. The same
should also govern the B & D setups, which are set by hand, at P band,
to set to the same RMS.
I shall take some time tomorrow and see if turning down the Pband requantizer
gain to 8 or 10 helps in practice (too late to try 4 band - it's gone).