[evlatests] EVLA Antenna power issues
Jim Jackson
jjackson at aoc.nrao.edu
Wed Feb 8 19:43:03 EST 2006
We did some more tests on Antenna 16 to try to isolate the strange power
issues we have been seeing. Rob and I finally managed to duplicate the
symptoms by interrupting AC power to the fire alarm panel. At that point,
Bob, Steve and Ken came out from the servo shop and determined that the
fire alarm system was not behaving the way they expected it to when AC
power was interrupted. Apparently, there is a short dropout (maybe about
1/2 second) that occurs in the panel's power supply when it switches from
AC power to backup batteries. This glitch is disrupting the time delay
relays in both the main AC contactor panel and the 48VDC shunt trip
circuit. The servo guys came up with a straight forward modification to
the fire alarm panel to correct the problem and will be installing it shortly.
The second part of this was the mystery of why the DTS modules were failing
when this occurred. Tests have now confirmed that the 5 volt regulator
powering the MIB and PSC board in that module shuts down when a short
glitch occurs on it's input. After a conversation with Doug and Kerry, we
believe that adding a capacitor across the 48 volt input to this regulator
will solve the problem. This simple modification will be performed during
the next round of upgrades already planned for these modules sometime in
the next couple of months.
Cheers,
Jim
More information about the evlatests