A quick test was run this afternoon to test whether the 180/90
degree holography phase effect is really a function of bandwidth (not
expected), or is actually a function of the SSLO.
Short Answer: It's not the BW, but the SSLO which determines
whether these big jumps occurs.
Method:
Three quick in-beam holography blocks, each consisting of three
11-step back-and-forth holography scans were set up:
#1) 1485/1385 (AC/BD) MHz, with 25 MHz bandwidth. Past experience
shows we expect no major phase jumps.
#2) 1485/1385 MHz, with 12.5 MHz bandwidth. We expect 180 degree
phase jumps.
#3) 1378.75/1478.75 MHz, with 12.5 MHz bandwidth. This frequency is
selected to have the same LO tuning as block #1.
Results:
#1: No major phase jumps. Two `10-second' jumps are seen, each on
the first point following the start of a holographic scan.
#2: 180-degree phase jumps occured on the first two holographic
scans. The third scan was unaffected. In addition three `10-second'
variety phase jumps occurred,
#3: No major phase jumps. Two `10-second' jumps are seen, similar
to #1.
Thus, it's not the BW per se, but the actual tuning values, which
causes the major jumps.