[evlatests] L-Band sensitivity on 13, 14 and 16
Dan Mertely3
dmertely at aoc.nrao.edu
Wed Apr 26 14:56:11 EDT 2006
Rob: Do you know what frequency you had to tune
50 MHz off from to make the RFI go away?
-Mert
Rob Long wrote:
> I don't believe that there any differences at L vs C band through the
> LSC converter. We could look at this again on the antennas more
> closely, but I thought that Rick and Bob already checked Tsys through
> the T304 downconverter output.(?)
>
> The other thing I have noticed lately is a fair amount of RFI at L-band.
> When I use the maint script to set up L-band, I see badly distorted or
> completely indiscernible switched power at the T5 detector (using an
> o'scope.) It is not usually all 4 IFs, but it sometimes shows in either
> one or the other polarization. I have had to retune the L302s (most
> typically the L302-2) about 50 MHz to reduce the peak power, stop
> saturating the digitizers, and see reasonable switched power at the T5.
>
> Rob
>
> Ken Sowinski wrote:
>
>>
>>
>> On Tue, 25 Apr 2006, Walter Brisken wrote:
>>
>>>
>>> On Tue, 25 Apr 2006, Rick Perley wrote:
>>>
>>>> So where is the problem? Bob Hayward and I did efficiency and Tsys
>>>> tests at the front end on antenna 13 which clearly showed that G/T
>>>> should
>>>> be about the same as VLA -- perhaps at most reduced by 10%.
>>>> I list some possible explanations: a) High Tsys. b) Low efficiency.
>>>> c) Bad pointing. d) Defocussed. e) Subreflector seriously out of
>>>> position.
>>>
>>>
>>>
>>> I'd add to the list two other possibilities consistent with having a
>>> good receiver: f) out-of-subband power (RFI or internal) either
>>> causing clipping or inappropriate level setting, and g) low or
>>> unstable LO power at chosen tuning. Could g have an affect in the
>>> LSC converter that would not show up at C band?
>>
>>
>>
>> Rick has two consistent lines of evidence for his conclusion:
>> analysis of the correlation coefficient and sync detector
>> deflection between blank sky and known calibrator. It is hard
>> to see how g) would affect the SD deflection. f) is certainly
>> a possibility. I was going to add a similar statement to Rick's
>> list. I think a case can be made that the sampler/transition
>> module can degrade both the correlation coefficent and the SD.
>> Perhaps we need to try a bandpass filter in front of the sampler
>> again to see what happens to L band when the input to the sampler
>> is band limited to make it more well behaved.
>>
>> _______________________________________________
>> evlatests mailing list
>> evlatests at listmgr.cv.nrao.edu
>> http://listmgr.cv.nrao.edu/mailman/listinfo/evlatests
>
>
> _______________________________________________
> evlatests mailing list
> evlatests at listmgr.cv.nrao.edu
> http://listmgr.cv.nrao.edu/mailman/listinfo/evlatests
More information about the evlatests
mailing list