You have already received advice. I do not know all of it.
to deal with your situation you need to
1. Run CLCOR on the the cal source correcting its position
2. Run CALIB with DOCALIB = 2 so that the CLCOR is applied to the cal
source before any instrumental phase is found
3. Run CLCAL to transfer the SN table to CL table (now 3 I guess)
4. Then with DOCAL=2, run IMAGR on the source
Your description of using the model at the correct position in a
self-cal merges the 2 effects and treats both as instrumental phases -
so that the position correction phase of the cal source is applied to
the target source which is wrong.
Eric Greisen