[daip] How should I interpret this?

Eric Greisen egreisen at NRAO.EDU
Wed Jul 19 09:42:50 EDT 2000


Johan van der Walt writes:
 > I have started the editing and calibration of the data of a run we had
 > on the VLA in June. 

      This does sound odd - can you give me the VLA obs code for this
run?  And a little more about which source etc?

 > 
 > I have followed the procedure in section 4.3.1 (initial editing) and
 > noted that for a specific calibrator the values are round about 800 -
 > 900 for about the first 40 seconds on the source. For the remaining 80
 > seconds on the source the values given by LISTR are between 80 and 90,
 > ie 10 times smaller than during the first 40 seconds. This applies to
 > all baselines. It also seems to occur only for one specific calibrator
 > source but not for everytime when it was observed.
 > 
 > Is this bad data? 
 > 

     The "fluxes" with no calibration should be roughly in 0.1 Jys
rather than Jys.  That should tell you whether the 800 or the 80 is
the more correct.


Eric



More information about the Daip mailing list